No, ISO 13485 does not cover directly the incoming sampling inspection. Requirement 7.4.3 is stated that the manufacturer must implement inspection activities based on the supplier evaluation results and proportional to the risk associated with the purchased product.
However, most manufacturers who have medical devices organized in lots use ISO 2859-1:2012 Sampling procedures for inspection by attributes — Part 1: Sampling schemes indexed by acceptance quality limit (AQL) for lot-by-lot inspection.
I agreed what you said. My question is related to ISO 13485 8.2.6 Monitoring and measurement of product.
In that clause, it is mentioned that "The organization shall monitor and measure the characteristics of the product to verify that product requirements have been met."
Is that same as incoming goods? If yes then, we do not need to write process description if i already wrote it for incoming goods inspection?
Thanks!!!
No this is not referring for the incoming goods, rather for final product. The only way how this can be the same if you are purchasing the final product that you than put on the market and not doing anything with it.