Our intension is to provide certified values that are traceable to an International Standard. The workflow is the following: 1. External lab is certifying the power of a set of lenses that cover the calibration range. This lab provides (according to their certificate) NIST traceable values 2. We purchase a calibration target another provider that also confirm that this target is NIST Traceable 3. We calibrate our lab metrology device with the certified devices (mentioned above) The questions are as follows: 1. If we measure another set of calibration lenses in house with our now calibrated lab metrology device, can we claim the power of these lenses are NIST Traceable? (no longer requiring the use of another external provider) 2. If so, my assumption here is that we would need to confirm through documentation/audit of the original external lab and target provider to confirm that the chain is unbroken - all certificates/claims are valid etc. Please can you share your thoughts regarding this?
0 0

Assign topic to the user

ISO 17025 DOCUMENTATION TOOLKIT

Step-by-step implementation for smaller companies.

ISO 17025 DOCUMENTATION TOOLKIT

Step-by-step implementation for smaller companies.

Expert
Tracey Evans Apr 14, 2026

Hi, I am basing my answer on the assumption you are a testing, not a calibration laboratory.

You asked

1. If we measure another set of calibration lenses in house with our now calibrated lab metrology device, can we claim the power of these lenses are NIST Traceable? (no longer requiring the use of another external provider)

Yes, provided that you establish and maintain metrological traceability through an unbroken, documented calibration chain, and evaluate the associated measurement uncertainty. According to ISO 17025, metrological traceability requires you to link measurement results to a reference (e.g., NIST) through a documented, unbroken chain of calibrations, each contributing to measurement uncertainty.

Therefore, if your metrology device is calibrated using traceable standards, the calibration covers the required range, and measurement uncertainty is evaluated and controlled, then the values you assign to lenses can be claimed as traceable to NIST.

You also asked

2. If so, my assumption here is that we would need to confirm through documentation/audit of the original external lab and target provider to confirm that the chain is unbroken - all certificates/claims are valid etc.

What you need to do is ensure the competence and suitability of external providers, but you are not required to audit the full traceability chain back to NIST. This is typically achieved by using accredited laboratories or evaluating non-accredited providers based on technical competence.  Yes, check that they are accredited or competent (through proven records) and that their certificates include traceability and report uncertainty, and that they retain documented evidence supporting the traceability claim.

Quote
0 1
smc Apr 15, 2026

Thanks Tracey. If we are planning on being a Testing & calibration laboratory (ISO17025) and would like to issue calibration certificates (for a set of calibration lenses) for use by others as part of the process above, what amendments need to be considered? (we will perform uncertainty budgets etc.)

Quote
0 0
Expert
Tracey Evans Apr 15, 2026

Hi you asked

We are planning on being a Testing & calibration laboratory (ISO17025) and would like to issue calibration certificates (for a set of calibration lenses) for use by others as part of the process above, what amendments need to be considered? (we will perform uncertainty budgets etc.)

You can issue traceable calibration certificates, to be used by testing laboratories provided you demonstrate full technical competence, control, and properly evaluated uncertainty in line with ISO 17025 for your metrology scope.

This includes, besides ensuring an unbroken traceability chain to NIST

  • Defining your calibration capability (range and uncertainty) 
    Specify the lens power range covered and the uncertainty you can achieve
  • Using validated or verified calibration methods
    Ensure the method is fit for purpose and performs consistently.
  • Developing and reporting measurement uncertainty
    Establish an uncertainty budget (including standards, equipment, environment, and repeatability) and report it with results.
  • Maintaining calibrated and controlled equipment
    Apply a calibration programme with periodic recalibration and checks.
  • Implementing ongoing checks for the validity of results
    Use check standards or repeat measurements to confirm performance.
  • Issuing compliant calibration certificates
    Here you will be guided by your accreditation body. Include results, uncertainty, traceability, and relevant conditions. to meet the requirements of ISO 17025 for reporting calibration results.