Does this mean that if you have mass pieces that are verified on a balance right after calibration of the balance, this is insufficient to prove 'fit for purpose'?
No that would not be sufficient. To meet the requirement of clause 6.4.5 and 6.4.6, as well as 6.5 such mass pieces should be calibrated (not verified) by a “competent” laboratory. i.e either by a calibration laboratory accredited to ISO 17025 for mass, or at least one which meets the requirements to perfom and report the calibration competently according to ISO 17025.
You also asked a number of questions regarding reporting. Note that ISO 17025 states that all must be included unless you have a valid reason for not. This tells you if you have considered the risk and are confident the results are reported suitably (not ambiguous) then you can justify a simplified report.
7.8.2.1 e) - If this is an inhouse client, then a simplified report is suitable – as long as the sample is traceable and you have evidence of who the report was provided to.
7.8.2.1 o) The way the authorisation takes place is not the issue, as long as it is traceble as to who authorised the report.
7..8.1.3 No not necessarily in a report, but the information must be available and traceble. Imagine there is a query – all information should be protected (from change) and be available on looking, for example, within a LIMS, database or a record.