MDR article 120 significant changes

  Quote
Guest
Guest user Created:   Aug 05, 2022 Last commented:   Aug 08, 2022

MDR article 120 significant changes

I need your support on an issue, we changed the contact sterilization supplier of our class I sterile product.
Is this significant change according to MDR article 120?
I need to do a decision tree evaluation on this,
I need information on this topic.

0 0

Assign topic to the user

ISO 13485 & EU MDR INTEGRATED DOCUMENTATION TOOLKIT

Step-by-step implementation for smaller companies.

ISO 13485 & EU MDR INTEGRATED DOCUMENTATION TOOLKIT

Step-by-step implementation for smaller companies.

Expert
Kristina Zvonar Brkić Aug 08, 2022

Yes, it is, because you need to prove that the new company that provides sterilization is just as good as the previous one (so that it has a valid ISO 13485 certificate for its type of sterilization, that you have sterilization validation results with the new provider, that you have done biological indicator tests with the new provider, etc.). Also, in your risk analysis, you should check whether any new risks have appeared with the change of sterilization provider: for example, how long is the transport from you to the new provider, is it longer or shorter, has the responsibility for transport changed.

Quote
0 0