Requirement for Post Market Surveillance
We manufacturer Class 1 medical devices and I am wondering what we are required for the Post Market Surveillance procedure?
Assign topic to the user
According to Article 83 - Post-market surveillance system of the manufacturer - is stated that for each device, manufacturers shall plan, establish, document, implement, maintain and update a post-market surveillance system in a manner that is proportionate to the risk class and appropriate for the type of device. Then, in Article 85 - Post-market surveillance report - Manufacturers of class I devices shall prepare a post-market surveillance report summarising the results and conclusions of the analyses of the post-market surveillance data gathered as a result of the post-market surveillance plan.
Therefore, class I medical devices are not exempt from the post-market surveillance system.
More information on the Post-market surveillance system you can find on the following link:
- What are the post-market surveillance requirements in the MDR? https://advisera.com/13485academy/blog/2021/04/29/what-are-the-post-market-surveillance-requirements-in-the-mdr/