i everyone,
We received a minor nonconformity for point 7.1.5.3.1.d of the standard. It states:
The laboratory shall specify and implement, as a minimum, requirements for:
d) capability to perform these services correctly, traceable to the relevant process standard (such as ASTM, EN, etc.); when no national or international standard(s) is available, the organization shall define and implement a methodology to verify measurement system capability;
We have an internally manufactured weighing machine that uses load cells. The auditor did not find any normative reference in our calibration procedure. I am not challenging this nonconformity, but I would like to get your advice to avoid recurrence.
What do you do when there are no national or international standards?
The standard says that a methodology must be defined and implemented to verify the capability of the measurement system. Do you have an example of such a method? Is this redundant with the MSA requirement? I asked one of our consultants, who advised me to calculate a Cgk. But does that really make sense?
In short, I would like to understand how you address this point of the standard and what you do when no standards exist especially regarding capability analysis method.
Many thanks in advance.
IATF 16949 eğitimlerini alacağımız yerler Certification Body(Bureau Veritas, SGS, BSI) ya da Lead Auditor olmak zorunda mıdır? AIAG CORE TOOLS eğitimleri için de aynı zorunlulukar geçerli midir? Danışmanlık ve belgelendirme adı altında piyasada birçok kuruluş var olup bu eğitimleri veriyor. Buralardan alınan sertifikalar geçersiz mi olur? En temel haliyle IATF 16949 Temel Bilgilendirme, İç Denetçi ve Core Tools eğitim sertifikalarını veren yerlerde hangi yeterlilikler olmalıdır?
Hi experienced men
Could you please share me the CQI17 implement experiences, including: Checklist template, Procedure,...?
I'm learning to prepare it for IATF audit and certification.
Thanks and Best regards
Is FMEA new style compulsary to implement for new projects, previously set up in FMEA <2016 but now transferred to a different company. Or can we continue on the previous set up old style FMEA and copy this after relocation of the product?
I would like to know what would be on a list of deliverables for an NPI project in an IATF-certified organization
According to IATF1649, if there are documents used to record test results that are used every day by Production Employees should these documents be Controlled Documents?
I was wondering if there's any specific way to identify controlled updated prints/documents per IATF ? Is using a stamp permitted?